CHATPRIVACYDONATELOGINREGISTER
DMT-Nexus
FAQWIKIHEALTH & SAFETYARTATTITUDEACTIVE TOPICS
pH meter problems again! HELP! Options
 
69ron
#1 Posted : 5/30/2009 10:08:35 PM

DMT-Nexus member


Posts: 5826
Joined: 09-Jun-2008
Last visit: 08-Sep-2010
Location: USA
I'm having pH meter problems again!

I hate this. All the pH meters I have seem to get the same problem after a while. They become extremely erratic.

I have gone through dozens now. All end with the same problem. Here's the symptoms:

Read: pH meter shows 14, then 0, then 3, the 8, then 2, etc., etc., etc.

Shake pH meter: pH meter looks fine for a minute or two, then back to showing erratic numbers.

What causes this problem?

Why does shaking the meter temporarily fix the problem?

How can I fix this or do I just have to buy another one?

These are all very expensive Hanna meters by the way in the price range of $100-$500 that have developed this problem.

I've had this same problem with so many now, I'm so sick of it.
You may remember me as 69Ron. I was suspended years ago for selling bunk products under false pretenses. I try to sneak back from time to time under different names, but unfortunately, the moderators of the DMT-Nexus are infinitely smarter than I am.

If you see me at the waterpark, please say hello. I'll be the delusional 50 something in the American flag Speedo, oiling up his monster guns while responding to imaginary requests for selfies from invisible teenage girls.
 

Live plants. Sustainable, ethically sourced, native American owned.
 
WSaged
#2 Posted : 5/30/2009 11:51:08 PM

DMT-Nexus member

Senior Member

Posts: 1813
Joined: 18-Jan-2008
Last visit: 17-Oct-2013
Location: Heart of the Sun
Is it the kind of meter where you can change out the pH detection element?
I don't think that's the correct term for it, but the head on the meter, some of the expensive ones can have the head replaced.
Have you tried that? Probably a dumb ??, but that's where I'd start.

I've also read somewhere...this was a while ago, when I was learning to use a pH meter, so I could be off a bit here...that you can either store the removable heads in a buffer solution to keep them true.(?)
Or soak a pH meter in buffer solution over night to clean it.
(please check those suggestions out before doing them, to make sure I remember correctly!!)


Cheers,

WS
All posts are fictional short stories depicting the adventures of WSaged!! None of these events have actually happened and any resemblance to any real persons or incidents is totally coincidence!!!!!!!!!!!!
 
SWIMfriend
#3 Posted : 5/30/2009 11:58:30 PM

DMT-Nexus member

Senior Member

Posts: 1695
Joined: 04-May-2009
Last visit: 11-Jul-2020
Location: US
Problems with pH meters are invariably the probe electrode end. Of course they need to be nicely cleaned when used (distilled water). They often need to be re-calibrated; and they often need to be replaced. Even in university labs, pH meters are a source of frustration.

Wikipedia as an informative and short article on the pleasures and miseries of the digital pH meter.
 
69ron
#4 Posted : 5/31/2009 5:03:07 AM

DMT-Nexus member


Posts: 5826
Joined: 09-Jun-2008
Last visit: 08-Sep-2010
Location: USA
warrensaged wrote:
Is it the kind of meter where you can change out the pH detection element?
I don't think that's the correct term for it, but the head on the meter, some of the expensive ones can have the head replaced.
Have you tried that? Probably a dumb ??, but that's where I'd start.

I've also read somewhere...this was a while ago, when I was learning to use a pH meter, so I could be off a bit here...that you can either store the removable heads in a buffer solution to keep them true.(?)
Or soak a pH meter in buffer solution over night to clean it.
(please check those suggestions out before doing them, to make sure I remember correctly!!)


Cheers,

WS


Yeah it's the removable kind. The correct term is probe. The probes keep going bad and I keep having to replace them. The probes themselves are pricey. I’ve tried all kinds from all different price ranges from el-cheapo to top of the line and all of them go bad in the same way. The other part works fine and has worked for many years now, it’s just the probes that are going bad.

The symptom is the same, they go from working perfectly to suddenly going bonkers. For example, I measure a solution at pH 5.4, it works for a while. I keep mixing the solution with the probe in the solution and then suddenly it goes bonkers giving readings that are completely random from one end of the pH spectrum to the next. Cleaning doesn’t solve the problem, but vigorously shaking the probe does, but only for a few minutes and it then goes back to it’s crazy readings again. It’s almost like there's a short in the wire in the probe and shaking it causes it to reconnect.

This has happened to every single pH probe.

I’ve had some show this problem after a week of use! And others last for a few years and then start showing this problem.

Is there something that can cause this that maybe I’m doing wrong with the meter? What can cause it to go from reading fine, to being random, and then back to being fine after vigorously shaking it? Sometimes shaking it works for a few days, other times just a few seconds and its all bonkers again.

Does anyone else have any experience with this problem and know of any solutions or does this just mean the probe it end of life and beyond fixing?

These probes are the kind that are refillable. Would refilling it make any difference?
You may remember me as 69Ron. I was suspended years ago for selling bunk products under false pretenses. I try to sneak back from time to time under different names, but unfortunately, the moderators of the DMT-Nexus are infinitely smarter than I am.

If you see me at the waterpark, please say hello. I'll be the delusional 50 something in the American flag Speedo, oiling up his monster guns while responding to imaginary requests for selfies from invisible teenage girls.
 
SWIMfriend
#5 Posted : 5/31/2009 5:10:15 AM

DMT-Nexus member

Senior Member

Posts: 1695
Joined: 04-May-2009
Last visit: 11-Jul-2020
Location: US
69ron wrote:
Is there something that can cause this that maybe I’m doing wrong with the meter?


Probes really are supposed to be used in a "dip/clean/return to storage solution" manner. They aren't supposed to stay in an unknown solution for a long time. The storage solution is typically supposed to be at pH 3, I believe. If you were to leave a probe in a solution far from that for a long time, it would likely at least lose calibration.

(the above was true, at least, ten years ago. I haven't really used pH meters much since then).
 
Xstacy
#6 Posted : 5/31/2009 5:28:44 AM
DMT-Nexus member


Posts: 281
Joined: 21-Apr-2009
Last visit: 12-Nov-2012
Location: US
Dont use them, go STB.
All illegal narcotics are medicinal. Boredom is a disease worse than cancer. Drugs cure it, with little or no side effects if used as directed - Doug Stanhope.

Rightful liberty is unobstructed action according to our will within limits drawn around us by the equal rights of others. I do not add 'within the limits of the law' because law is often but the tyrant's will, and always so when it violates the rights of the individual. - Thomas Jefferson
 
69ron
#7 Posted : 5/31/2009 5:40:44 AM

DMT-Nexus member


Posts: 5826
Joined: 09-Jun-2008
Last visit: 08-Sep-2010
Location: USA
I clean them and do all the normal maintenance they need. They are stored in proper storage solution designed specifically for them.

The most they're in a solution that I'm reading the pH of is maybe a few minutes and then they are returned back to their storage solution.

This problem has nothing to do with calibration. They are properly calibrated and go back to being properly calibrated after a vigorous shake.
You may remember me as 69Ron. I was suspended years ago for selling bunk products under false pretenses. I try to sneak back from time to time under different names, but unfortunately, the moderators of the DMT-Nexus are infinitely smarter than I am.

If you see me at the waterpark, please say hello. I'll be the delusional 50 something in the American flag Speedo, oiling up his monster guns while responding to imaginary requests for selfies from invisible teenage girls.
 
SWIMfriend
#8 Posted : 5/31/2009 5:59:45 AM

DMT-Nexus member

Senior Member

Posts: 1695
Joined: 04-May-2009
Last visit: 11-Jul-2020
Location: US
69ron wrote:
I clean them and do all the normal maintenance they need. They are stored in proper storage solution designed specifically for them.

The most they're in a solution that I'm reading the pH of is maybe a few minutes and then they are returned back to their storage solution.

This problem has nothing to do with calibration. They are properly calibrated and go back to being properly calibrated after a vigorous shake.


I can only say that they've been somewhat of a pain in many labs. Most people just accept that the probes have to get changed fairly regularly.

OK, one thought (don't hold me to this). Don't the probes actually have a tiny hole in them (or some sort of semi-permeable area)? Maybe you're getting some crystalization (or some other clog--minute particles of bark), and when you shake it the hole clears up. Maybe they all get better if you shake them; I wouldn't know because I've only ever seen people replace them immediately when they go whacky.
 
SWIMfriend
#9 Posted : 5/31/2009 6:11:45 AM

DMT-Nexus member

Senior Member

Posts: 1695
Joined: 04-May-2009
Last visit: 11-Jul-2020
Location: US
This little article says you should expect to change the probe every 1 to 6 months.
 
geeg30
#10 Posted : 5/31/2009 10:58:33 PM

DMT-Nexus member


Posts: 440
Joined: 08-Sep-2008
Last visit: 07-Sep-2019
Location: top left corner of a £20 note
SWIM uses a cheapo hydroponics digital meter that he has had for about 4 years and it still gives good results, just needs recalibrating every now and again. Swim also found that lab meters were a bit erratic when he was at University so therefore reasons this is due to the tolerances of the lab equipment i.e a small scratch on the probe would send the meter way off.
Here you!!! Gonnaenodaethat

"Iceberg???? - What Iceberg????"
 
WSaged
#11 Posted : 6/1/2009 12:31:24 AM

DMT-Nexus member

Senior Member

Posts: 1813
Joined: 18-Jan-2008
Last visit: 17-Oct-2013
Location: Heart of the Sun
Ya know, my $30 generic pH meter has been pretty damn accurate for over a year now.

Quote:
Don't the probes actually have a tiny hole in them (or some sort of semi-permeable area)?


Is that how it works??
I must have studied that little glass tip for over an hour one time, trying to figure out how pH meters read anything through the glass bulb!!

Do you know anything else about how they use a glass bulb to read pH levels?
Kind of interesting!

WS
All posts are fictional short stories depicting the adventures of WSaged!! None of these events have actually happened and any resemblance to any real persons or incidents is totally coincidence!!!!!!!!!!!!
 
SWIMfriend
#12 Posted : 6/1/2009 12:37:46 AM

DMT-Nexus member

Senior Member

Posts: 1695
Joined: 04-May-2009
Last visit: 11-Jul-2020
Location: US
Wikipedia has a quick and easy article about pH meters, and a link in that article to "glass electrodes," which explains how they work. Yeah, they are permeable in one area.

I've seen pH probes get KCL precipitated on them or around them--in their "storage solution." Seems to me if you got some gunk like that on the permeable area (usually a bit above the actual tip--on the side), then it wouldn't work--and vigorously shaking MIGHT temporarily cure the problem. That's total speculation, though.
 
benzyme
#13 Posted : 6/1/2009 12:58:18 AM

analytical chemist

Moderator | Skills: Analytical equipment, Chemical master expertExtreme Chemical expert | Skills: Analytical equipment, Chemical master expertChemical expert | Skills: Analytical equipment, Chemical master expertSenior Member | Skills: Analytical equipment, Chemical master expert

Posts: 7463
Joined: 21-May-2008
Last visit: 03-Mar-2024
Location: the lab
also make sure they're used away from ESD sources. the electrodes are very sensitive to static

swim also has a Hanna portable meter, been working well for about three years now.
"Nothing is true, everything is permitted." ~ hassan i sabbah
"Experiments are the only means of attaining knowledge at our disposal. The rest is poetry, imagination." -Max Planck
 
D00R5
#14 Posted : 6/1/2009 1:39:29 AM

DMT-Nexus member


Posts: 7
Joined: 31-May-2009
Last visit: 23-Apr-2013
Solute and respect to DMT-Nexus Community! This place and people have been a tremendous source of knowledge and inspiration for me in search for an “ultimate truth” we all seek. For past few years I was devoted lurker of this site. I am grateful for everything I found here but feel kind of guilty for not even showing up here once to say thanks or give something back.

Quote:
Problems with pH meters are invariably the probe electrode end.

“SWIMfriend” s remark is very true. A lot of careful maintenance is needed on electrode probe part of PH meter units. The PH meter itself does not need to be calibrated often. It is only done by labs maybe like once a year when they send the units back to manufacturer for recertification. For the type of experimentation we do it is not that critical at all and of course far from being economically feasible. The obviously easiest way to check Ph meter for at least a relevant accuracy is to plug in a calibrated, working electrode probe of right kind. That is why the best technique (when affordable-) is to have 2 channel PH meter with 2 probes connected which enables you to notice obvious deviations between the two in readings of the same sample. If there is no spare electrode available you can run a simple and generally reliable meter test. From Ph mode reading turn to MV mode. Disconnect the probe and short out the meter with shorting strap. It should read 0.0mv which is = pH 7.0. If it is out of range for more than -+2 the meter should be electronically calibrated by manufacturer. Too bad. If electrode is suspected to be faulty fallowing can be done. Try measure MV with pH buffer 7.0. The reading should be 0.0mv no more than +-20mv. Also with pH 4.0 or 7.0 it should read between 160-180mv. If off by more than +-12 of 177mv – too bad. The probe should be replaced or you can try reconditioning. Reconditioning can include such techniques as unblocking clogged junctions, checking for its crystallization and changing filling solution, cleaning glass membrane. There are tons of articles on how to on the net. One of the good sources is fishersci.com look up the section with technical documentation in pdf on electrode maintenance. OR coleparmer.com/techinfo/TechInfo_ByProduct.asp?ProdID=3527 (Product Maintenance section). In your case Ron, I would simply try to shake out the solution and refill it (which I bet you already tried). Accidentally scratched glass membrane would also contribute to your collection of dead probes. Avoid putting any kind of electric devices close to ph meter and especially electrode cable. Regular combination probes do not tolerate organic solutions very well – double junction electrodes or at least non-refillable (cheaper) are probably best for our kind of work. And unfortunately, yes there is always a place for a mystery with this kind of highly sensitive apparatus when you not operating out of state of the modern art laboratory.-) All the best everyone.
 
AI
#15 Posted : 7/13/2009 7:51:48 PM
DMT-Nexus member


Posts: 8
Joined: 09-Jun-2008
Last visit: 07-Aug-2009
69ron wrote:

How can I fix this or do I just have to buy another one?

These are all very expensive Hanna meters by the way in the price range of $100-$500 that have developed this problem.

I've had this same problem with so many now, I'm so sick of it.


It's generally accepted in the cannabis growing community that pH meters (and especially the probes themselves) are fidgety. While the meters most growers use is probably slightly less expensive, a bunch of them do splurge on the more expensive meters (including meters by Hanna). Complaints are frequent of having to change probes often and of meters simply going haywire and not taking accurate readings. I have found that storing my probe in some 4.01 or 7.01 calibration solution, and frequently rinsing it between readings with either of the solutions helps to keep it operational. I also have to calibrate it when it seems to be giving unexpected results. But, in general, I've found it's easier to use pH paper for my purposes.
 
noone
#16 Posted : 10/12/2009 4:10:41 PM
DMT-Nexus member


Posts: 15
Joined: 19-Aug-2009
Last visit: 04-Jan-2010
I had my meter freak out on the second use. Turns our when measuring high PH (in my case 13.4) the probe starts to react at levels over 10. Second time I used it would only range from -5 - +2.0. Rinsed it out really good and shook it and the next day it worked 0.1 out of calibration. I think I am going to return it and get papers...
 
 
Users browsing this forum
Guest

DMT-Nexus theme created by The Traveler
This page was generated in 0.057 seconds.